This is likely due to the hold vacuum being set too high. In general, the hold pressure regulator controls both the hold vacuum as well as the clear-hold pressure. The higher the holding vacuum, the higher clear-hold pressure will be.
Press the CLRH button to push the cell out of the holding pipette. If CLRH shoots the cell too far away, reduce the hold pressure.